cmackenzie
Joined: 16 Dec 2011 Posts: 2 Location: Canada
|
Posted: Thu Jan 05, 2012 5:08 pm Post subject: iontoporesis currents are not staying constant |
|
|
I'm using the Idromed PS 5 and when I put the current to fully on, it shows 25 mA (sometimes even 26 mA). When I place my feet in the tubs, the current drops and fluctuates between 19 and 22 mA. The longer I keep my feet in the higher it typically will go. Essentially, I am almost always unable to get to 25 mA.
Should I be concerned about this, is my device working properly? |
|
| Back to top |
|
 |
support Site Admin
Joined: 20 Mar 2008 Posts: 1258
|
Posted: Thu Jan 05, 2012 6:32 pm Post subject: |
|
|
Hello,
Don't worry, this is probably due to the high body resistance to the current. Have you maybe used cream soap before treatments? Please read trough following link teaching you ībout skin care for iontophoreis and let us know!
Also the bodys resistant is floating whilst a iontoporesis treatment.
When you are "near the border" of the mA measurement, the display can show you sometimes a switch between 2 direct benath 2 neigbour numbers. _________________ You are welcome...
Support Staff
We are reachable from 9am to 5pm !Board time!
mail: info-at-iontophoresis-device.com |
|